Online Blind Calibration of Non-Uniform Photodetectors: Application to Endomicroscopy - Inria - Institut national de recherche en sciences et technologies du numérique Accéder directement au contenu
Nicolas Savoire
  • Fonction : Auteur
  • PersonId : 841864
Barbara André
  • Fonction : Auteur
  • PersonId : 902248
We present an original method for the online blind calibration of non-uniform photodetectors. The disparity of the detectors may arise from both irregular spatial arrangement and distinct slowly timevarying photometric transfer functions. As natural images are mostly continuous, the signal collected by neighboring detectors is strongly correlated over time. The core idea of our method is to translate the calibration problem into relative pairwise calibrations between neighboring detectors followed by the regularized inversion of a system akin to gradientbased surface recovery. From our blind calibration procedure, we design an online blind calibration pipeline compatible with clinical practice. On- line blind calibration is proved to be statistically better than standard offline calibration for reconstructing endomicroscopy sequences.
MICCAI12-CamReady.pdf (247.18 Ko) Télécharger le fichier
hal-00813864 , version 1 (03-05-2013)
  • HAL Id : hal-00813864 , version 1
Nicolas Savoire, Barbara André, Tom Vercauteren. Online Blind Calibration of Non-Uniform Photodetectors: Application to Endomicroscopy. Proceedings of Medical Image Computing and Computer Assisted Intervention 2012 (MICCAI), 2012, Unknown. ⟨hal-00813864⟩
Collections
TDS-MACS
61 Consultations
84 Téléchargements