A Novel Temporal Calibration Method for 3D Ultrasound - Inria - Institut national de recherche en sciences et technologies du numérique Accéder directement au contenu
Article Dans Une Revue IEEE Transactions on Medical Imaging Année : 2006
This paper examines a novel approach for temporal calibration of a 3D freehand ultrasound (US) system. A localization system fixed on the probe gives the position and orientation of the probe. For quantitative use, calibration is needed to correctly localize a B-scan in 4D (3D+t) space. Temporal latency estimation is defined in a general robust formulation using no specific probe motion constraints. Experiments were performed on synthetic and real data using a 3D freehand ultrasound system. The achieved precision is lower than the image acquisition rate ($40ms$). A validation study using a calibration phantom has been performed to evaluate the influence of incorrect latency estimation on the 3D reconstruction procedure. We showed that for latency estimation errors less than $40ms$, the 3D reconstruction errors are negligible for volume estimation.
TMI-3DUS-TemporalCalibration.pdf (426.76 Ko) Télécharger le fichier
inria-00001215 , version 1 (05-04-2006)
  • HAL Id : inria-00001215 , version 1
François Rousseau, Pierre Hellier, Christian Barillot. A Novel Temporal Calibration Method for 3D Ultrasound. IEEE Transactions on Medical Imaging, 2006. ⟨inria-00001215⟩
206 Consultations
435 Téléchargements