Analytical assessment and performance of the 0/3h algorithm with novel high sensitivity cardiac troponin I
Background: Recently, Ortho Clinical Diagnostics have launched a high sensitivity cardiac troponin I assay adapted on Vitros3600 / 5600. We aimed at evaluating the analytical and clinical performances using the 0/3-hour algorithm, of the Ortho hs-cTnI assay on the Vitros 3600® (Ortho Clinical Diagnostics, Illkirch, France) analyzer with comparison to Roche hs-cTnT assay on the Cobas8000/e801® module (Roche diagnostics, Meylan, France) and the Abbott STAT hs-cTnI assay on the Alinity i® (Abbott, Rungis, France) analyzer.
Methods: Imprecision studies, linearity, limit of detection, and the interference to hemolysis were performed for Ortho hs-cTnI assay. The concordance study was based on results of troponin obtained from 160 patients with chest pain to the emergency department. Testing was performed simultaneously measuring the Roche hs-cTnT and the Ortho hs-cTnI, then on remaining sample the concentration of Abbott hs-cTnI (n = 150). We applied the 0/3h algorithm to rule out/rule in, in acute myocardial infarction.
Results: Analytical performances were acceptable and in accordance with manufacturer’s data. For late presenters 100, 92.8 and 88.8% patients could be rule-out with Roche, Ortho and Abbott assay, respectively with one NSTEMI missed with both hs-TnI assays. Applying the hs-cTn cut-offs from 0/3-hour algorithm, 107 (66.8%) could be rule out with Roche hs-cTnT with no AMI missed (sensitivity 100% [95%CI : 90.5 to 100] and NPV 100%); 89 with Ortho hs-cTnI (sensitivity 97.3% [95%CI : 85.8–99.9] and NPV 98.8% [95%CI :92.7–99.8]); and 61 with Abbott hs-cTnI (sensitivity 97% [95%CI : 84.2–99.9] and NPV 98.4% [95%CI : 89.8–99.7]). For rule-in, 23.7% (n = 37) from the total population would be designated in this group, with a specificity of 86.9% (95%CI: 79.7–92.4) and PPV of 69.8% (95%CI: 59.4–78.5) vs specificity of 72.3% (95%CI:63.5–80.0) and PPV of 51.4% (95%CI: 44.1–58.2) with Roche hs-cTnT vs Ortho hs-cTnI respectively; and with Abbott, 33 patients had constitued this group with a specificity of 52.1% (95%CI:42.7–61.4) and PPV of 36.4% (95%CI: 32.0–41.0).
Conclusion: In conclusion, according to ESC guidelines, our results indicate that the Ortho hs-cTnI assay for the Vitros 3600 instrument is a method that may be used in the clinical practice using 0/3-hour algorithm.