Article Dans Une Revue IEEE Transactions on Pattern Analysis and Machine Intelligence Année : 1996
We present a method for nonrigid motion analysis in time sequences of volume images (4D data). In this method, nonrigid motion of the deforming object contour is dynamically approximated by a physically-based deformable surface. In order to reduce the number of parameters describing the deformation, we make use of a modal analysis which provides a spatial smoothing of the surface. The deformation spectrum, which outlines the main excited modes, can be efficiently used for deformation comparison. Fourier analysis on time signals of the main deformation spectrum components provides a ternporal smoothing of the data. Thus a complex nonrigid deformation is described by only a few parameters: the main excited modes and the main Fourier harmonics. Therefore, 4D data can be analyzed in a very concise manner. The power and robustness of the approach is illustrated by various results on medical data. We believe that our method has important applications in automatic diagnosis of heart diseases and in motion compression.
NastarTPAMI1996.pdf (2.54 Mo) Télécharger le fichier
inria-00615057 , version 1 (17-08-2011)
Chahab Nastar, Nicholas Ayache. Frequency-based Non-rigid Motion Analysis: Application to Four Dimensional Medical Images. IEEE Transactions on Pattern Analysis and Machine Intelligence, 1996, 18 (11), pp.1067-1079. ⟨10.1109/34.544076⟩. ⟨inria-00615057⟩